Articles: respiratory-distress-syndrome.
-
To compare the accuracy of clinical observation and detailed respiratory recordings in identifying infants at high risk of developing pneumothoraces 10 infants, with idiopathic respiratory distress syndrome, were studied at three different ventilator rates. All infants with synchronous respiration at fast rates were correctly identified by clinical observation. ⋯ These criteria enabled correct identification of 'high risk' respiratory patterns on 15 (88%) of the 17 study occasions. These clinical criteria were then used as criteria for selective paralysis; no infant developed a pneumothorax during ventilation.
-
A new valveless ventilator, which uses an air jet to provide the driving force for positive pressure ventilation, was used on 13 newborn babies (10 of very low birthweight) who had severe respiratory disease. The ventilator differs from 'true' jet ventilators in that its driving gas does not take part in gas exchange. Functionally it is a pressure pre-set, time-cycled ventilator, whose performance is characterised by the rapid and precise maintenance of both inspiratory and expiratory airway opening pressure. ⋯ On the new ventilator, with the same settings, there was a dramatic and highly significant improvement within 20 to 30 minutes (mean values of arterial pCO2 were 6.45 kPa, pH 7.26, and inspired oxygen concentration 85.7%). This improvement was maintained. The new ventilator represents an important advance in the management of babies with severe respiratory failure.